Anesthesia Machine Checkout Recommendations
This checkout, or a reasonable equivalent, should be conducted before administration of anesthesia. These recommendations are only valid for an anesthesia system that conforms to current and relevant standards and includes an ascending bellows ventilator and at least the following monitors: capnograph, pulse oximeter, oxygen analyzer, respiratory volume monitor (spirometer), and breathing system pressure monitor with high and low pressure alarms.
• Verify backup ventilation equipment is available and functioning.*
• High Pressure System
o Check oxygen cylinder supply.*
o Check central pipeline supplies.*
• Low Pressure System
o Check initial status of low pressure system.*
o Perform leak check of machine low pressure system.*
o Turn on machine master switch and all other necessary electrical equipment.*
o Test flowmeters.*
• Adjust and check scavenging system.*
• Breathing System
o Calibrate O2 monitor.*
o Check initial status of breathing system.
o Perform leak check of the breathing system.
• Test manual and automatic ventilation systems and unidirectional valves.
• Check, calibrate, and/or set alarm limits of all monitors.
• Check final status of the machine.
o Vaporizers off.*
o APL valve open.
o Selector switch to 鈥淏ag鈥�.
o All flowmeters to zero.
o Patient suction level adequate.
o Breathing system ready to use.
*Note: If an anesthesia provider uses the same machine in successive cases, the steps denoted by an asterisk need not be repeated or may be abbreviated after the initial checkout.
麻醉机检查的建议
这一结帐,或一个合理的等效,前应进行麻醉管理。这些建议只适用于一个麻醉系统,符合当前和相关标准,包括一个提升波纹管呼吸至少有以下监测:capnograph,脉搏血氧仪,氧分析仪,监测呼吸量(呼吸),和呼吸系统压力监测与高,低报警。
•验证备份通风设备可以和运作。*
•高压力系统
阿检查氧气钢瓶供应。*
阿检查中央管道用品。*
•低压系统
阿检查初始状态的低气压系统。*
进行泄漏检查机低压系统。*
求你转向机上的主开关和其他一切必要的电气设备。*
测试流量计。*
•调整检查清除系统。*
•呼吸系统
阿校准氧监测。*
哦,初始状态的呼吸系统的检查。
进行泄漏检查呼吸系统。
•试验手动和自动通风系统及单向阀。
•检查,校准,及/或设置报警限制所有监视器。
•检查最终状态的机器。
哦,汽化器了。*
阿磷脂阀打开。
输出选择开关鈥�鈥淏银。
所有流量计零。
患者吸入水平足够的。
阿呼吸系统准备使用。
*注:如果一个麻醉提供商使用相同的机器在连续的情况下,将以星号表示不需要重复或可缩写后的初步检验。 |